Can i perform a calibration within TC-08 thermocouple data logger?
In other words, is there the possibility to insert a calibration of the thermocouple?
For example:
-) I buy a thermocouple
-) I found a table of value :
V Temperature
0 0
... ....
2 300
-) insert this table within datalogger so i perform the calibration and read the correct value of temperature
There are no calibration values that can be inserted into the USB-TC08 hardware, we at Pico can perform a Factory Calibration and Test to correct a unit that has drifted out of tolerance.
However it is possible to add scaling values within the PicoLog 6 software so that the readings are corrected in line with a calibration table of errors, if that is what you mean.
Thanks for the reply. I have another question. If i want to measure the humidity what type of datalogger should I use? It is possible to integrate it in TC-08?
You can use the single channel terminal board https://www.picotech.com/accessories/te ... inal-board with third party sensors that have ±5 V, ±500 mV and ±50 mV, or 4–20 mA current loop outputs. So if you can find a third party Humidity Sensor with one of these outputs that would be the recommended option.