Calibration TC-08 thermocouple data logger

Which product is right for your exact requirements
Post Reply
Lorenzo
User
User
Posts: 4
Joined: Mon Nov 07, 2022 12:18 pm

Calibration TC-08 thermocouple data logger

Post by Lorenzo »

Can i perform a calibration within TC-08 thermocouple data logger?
In other words, is there the possibility to insert a calibration of the thermocouple?
For example:
-) I buy a thermocouple
-) I found a table of value :
V Temperature
0 0
... ....
2 300
-) insert this table within datalogger so i perform the calibration and read the correct value of temperature

Thanks

Lorenzo
User
User
Posts: 4
Joined: Mon Nov 07, 2022 12:18 pm

Re: Calibration TC-08 thermocouple data logger

Post by Lorenzo »

If this datalogger can't perform this calibration, is there another hardware that can read thermocouples and perform this calibration?

Martyn
Site Admin
Site Admin
Posts: 4499
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Calibration TC-08 thermocouple data logger

Post by Martyn »

There are no calibration values that can be inserted into the USB-TC08 hardware, we at Pico can perform a Factory Calibration and Test to correct a unit that has drifted out of tolerance.

However it is possible to add scaling values within the PicoLog 6 software so that the readings are corrected in line with a calibration table of errors, if that is what you mean.
Martyn
Technical Support Manager

Lorenzo
User
User
Posts: 4
Joined: Mon Nov 07, 2022 12:18 pm

Re: Calibration TC-08 thermocouple data logger

Post by Lorenzo »

Thanks, it is possible to input this table of errors directly in the software? Can you share me a photo of this table in the software, please?

Martyn
Site Admin
Site Admin
Posts: 4499
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Calibration TC-08 thermocouple data logger

Post by Martyn »

Select a channel to setup
ChannelSetup.png
Click on Probe or Scaling
ProbesandScaling.png
Select table
TableScaling.png
and then add your calibration values
Martyn
Technical Support Manager

Lorenzo
User
User
Posts: 4
Joined: Mon Nov 07, 2022 12:18 pm

Re: Calibration TC-08 thermocouple data logger

Post by Lorenzo »

Thanks for the reply. I have another question. If i want to measure the humidity what type of datalogger should I use? It is possible to integrate it in TC-08?

Martyn
Site Admin
Site Admin
Posts: 4499
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Calibration TC-08 thermocouple data logger

Post by Martyn »

You can use the single channel terminal board https://www.picotech.com/accessories/te ... inal-board with third party sensors that have ±5 V, ±500 mV and ±50 mV, or 4–20 mA current loop outputs. So if you can find a third party Humidity Sensor with one of these outputs that would be the recommended option.
Martyn
Technical Support Manager