I am trying to measure the capacity of a battery pack by discharging it with a known resistor. For this I set a very large time/div (5000s/div) in order to able to capture about 10 hours. Unfortunately the software is unable to capture a voltage at such a large time/div setting. With shorter time/div setting (100ms/div) voltage looks good but with larger setting, the capturing does not work: It shows channel overrange. (See screenshot attached)
Normally during these 10 hours, voltage would drop from about 12.3v to 8v. So I choose a range of +/- 20v DC. At 8v the battery protection disconnects. In very some seldom cases after changing settings back and forth, I am able to capture without any issues. However most of the time I see the error.
My system: Thinkpad Carbon X1 Gen. 4, Ubuntu 18.04 LTS
Is there any workaround for this? Is this a known bug?
Any help would be appreciated.
Marc
Channel overrange on SW version 6.13.7.707
Re: Channel overrange on SW version 6.13.7.707
Do you get the same overrange indication at 1s/div, where the trace will still be streaming ?
If you leave it running at 5000s/div does the trace become reasonable after a few minutes ?
If you leave it running at 5000s/div does the trace become reasonable after a few minutes ?
Martyn
Technical Support Manager
Technical Support Manager
Re: Channel overrange on SW version 6.13.7.707
> Do you get the same overrange indication at 1s/div, where the trace will still be streaming ?
I will need to try, at which speed it start to happen.
> If you leave it running at 5000s/div does the trace become reasonable after a few minutes ?
No it does not. I waited at least more than 10 minutes and the result was during all the 10 minutes the channel overrange. Looking like a massive vertical block of data. Not even the slightest trace of data I would expect.
I will need to try, at which speed it start to happen.
> If you leave it running at 5000s/div does the trace become reasonable after a few minutes ?
No it does not. I waited at least more than 10 minutes and the result was during all the 10 minutes the channel overrange. Looking like a massive vertical block of data. Not even the slightest trace of data I would expect.
Re: Channel overrange on SW version 6.13.7.707
I have tried further:
Above and including "100 s/div" picoscope cannot show the signal properly. Faster speeds (50 s/div and faster) work without issues.
Steps to reproduce:
1. Add power supply to the picoscope (6402D)
2. Connect a 12v battery to Channel 1. (+ to CH.1 and - to shield)
3. Start picoscope from the command line.
4. Change the voltage/div from "Auto" to "±20 V"
5. Change the time/div from "5 ms/div" to "100 s/div"
Observe that it is running. Now instead of the expected 12v voltage a widening vertical column is shown (see below), with the text "Channel overrange" and when zooming in, the vertical column looks like some kind of modulated waveform (see below):
Above and including "100 s/div" picoscope cannot show the signal properly. Faster speeds (50 s/div and faster) work without issues.
Steps to reproduce:
1. Add power supply to the picoscope (6402D)
2. Connect a 12v battery to Channel 1. (+ to CH.1 and - to shield)
3. Start picoscope from the command line.
4. Change the voltage/div from "Auto" to "±20 V"
5. Change the time/div from "5 ms/div" to "100 s/div"
Observe that it is running. Now instead of the expected 12v voltage a widening vertical column is shown (see below), with the text "Channel overrange" and when zooming in, the vertical column looks like some kind of modulated waveform (see below):