Signal Loss when Zooming In on Y-Axis with PicoScope 3406DMSO

Post any questions you may have about our current range of oscilloscopes
Post Reply
Iver
Newbie
Posts: 0
Joined: Thu Mar 07, 2024 9:56 am

Signal Loss when Zooming In on Y-Axis with PicoScope 3406DMSO

Post by Iver »

I am trying to look at a very tiny returning signal from a
transducer. Parts of the signal starts to dissapear when zooming in on
the Y axis going from 500mV to 200mV (see attached pictures). Why is
this?
Attachments
image1.png
image2.png

Martyn
Site Admin
Site Admin
Posts: 4501
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Signal Loss when Zooming In on Y-Axis with PicoScope 3406DMSO

Post by Martyn »

Can you post the psdata files that match these images.
Martyn
Technical Support Manager

bennog
Advanced User
Advanced User
Posts: 208
Joined: Mon Nov 26, 2012 9:16 am
Location: Netherlands

Re: Signal Loss when Zooming In on Y-Axis with PicoScope 3406DMSO

Post by bennog »

you are using the wrong zoom button.
Use scroll wheel or the + icon not the [] icon.

Benno

Martyn
Site Admin
Site Admin
Posts: 4501
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Signal Loss when Zooming In on Y-Axis with PicoScope 3406DMSO

Post by Martyn »

That does look strange when changing the voltage range from +/-500mV to +/-200mV, it would be useful to know how high a voltage overrange there is, so changing from the +/-500mV range to +/-20V range would be interesting to see.
Martyn
Technical Support Manager

Iver
Newbie
Posts: 0
Joined: Thu Mar 07, 2024 9:56 am

Re: Signal Loss when Zooming In on Y-Axis with PicoScope 3406DMSO

Post by Iver »

PS data
Attachments
10V.psdata
(1019.26 KiB) Downloaded 32 times
500mV.psdata
(7.84 MiB) Downloaded 37 times
200mV.psdata
(3.01 MiB) Downloaded 30 times

Martyn
Site Admin
Site Admin
Posts: 4501
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Signal Loss when Zooming In on Y-Axis with PicoScope 3406DMSO

Post by Martyn »

Do you get the same effect if you try channels B to D ?

You will note that as you change from 500mV to 200mV a relay will click inside the unit as the input stage is switched to handle the smaller level signals. This will mean that the overvoltage near 10V signal you have at the input connector is significantly higher when it reaches the ADC chip on the 200mV range compared with the same signal on the 500mV range.
Martyn
Technical Support Manager

Iver
Newbie
Posts: 0
Joined: Thu Mar 07, 2024 9:56 am

Re: Signal Loss when Zooming In on Y-Axis with PicoScope 3406DMSO

Post by Iver »

Yes I am seeing the same effect on the other channels. All probes are set to 10x.
Attachments
4 probes.psdata
(6.95 MiB) Downloaded 35 times

Martyn
Site Admin
Site Admin
Posts: 4501
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Signal Loss when Zooming In on Y-Axis with PicoScope 3406DMSO

Post by Martyn »

If using 10x on the probe it is worth setting the software to the same, although this will only reflect in the correct voltages being displayed in the graph.
Attachments
ProbeScaling.png
Martyn
Technical Support Manager

Martyn
Site Admin
Site Admin
Posts: 4501
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Signal Loss when Zooming In on Y-Axis with PicoScope 3406DMSO

Post by Martyn »

The issue you are seeing is that on the 200mV and lower ranges the input amplifiers are saturating and they will take time to recover.

You can try using the graph zooming function on the 500mV or higher ranges to see more detail, potentially using some Resolution Enhancement on the Channel DSP tab to clean the signal a bit.
Martyn
Technical Support Manager

Iver
Newbie
Posts: 0
Joined: Thu Mar 07, 2024 9:56 am

Re: Signal Loss when Zooming In on Y-Axis with PicoScope 3406DMSO

Post by Iver »

I still ended up not seeing the returning signal I was looking for. But anyways thanks alot for the help! :D