Pico 2000A Timing and sampling settings

Post your LabVIEW discussions here
Post Reply
Joelvvdn
Newbie
Posts: 0
Joined: Tue Feb 20, 2024 3:30 pm

Pico 2000A Timing and sampling settings

Post by Joelvvdn »

Hello I need to measure ripple for a 12V adapter if I use Pico scope it shows a clear sin waveform. But When I try to acquire the same waveform I can't able to get the waveform.
I have attached the image of the pico7 screen when acquiring the voltage waveform.
In the time settings section in samples it is mentioned as 100ns/div, Sampled 125(1290), sample rate 125MS/s. how to set these options in LabVIEW. Please support on the above
Attachments
Screenshot 2024-02-20 210741.png

Martyn
Site Admin
Site Admin
Posts: 4501
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Pico 2000A Timing and sampling settings

Post by Martyn »

If you look at the Scope button on the top of the display you will see Samples 125 (1290), PicoScope 7 is only collecting 125 samples, but the softare is using sin(x)/x interpolation to bump this up to 1290 which is why you are seeing the ripple. If you are using the same sampling settings as PicoScope 7 in your LabVIEW application you will only have the raw samples, so would need to add interpolation.
Martyn
Technical Support Manager