Hello,

I would like to know if i can add an analog offset of -1V on channel A (or B) and what is in this case the input signal range?
Thanks

Andreas_
User
User
Posts: 6
Joined: Tue Aug 23, 2016 7:47 pm

Re: analog offset on picoscope 5000 series

Post by Andreas_ »

Hello,

the offset that you can apply is of course dependent on the selected range.
The odd thing with offsets is that it is not shown on the scaling.
So a -1V offset in the +/-2V range shows you a -1V value when the input is zero. (with effective -1 to 3V measurement range shown in a +/-2V display.

The alternative way is to define a custom specific probe.
E.g. for measuring the startup of a 31V power supply in the +/-20V range of the scope (1:1 input) you can define a custom specific probe showing a measurement range from -4 .. 36V (with a effective offset of -16V).