Peak to peak voltage errors

Post by Breesy31 »

I have recently purchased a Picoscope 2204A. I am trying to measure a ripple voltage on a DC rectifier circuit. I am measuring mV on the output of a LM7805 regulator but before even connecting to the regulated voltage my picoscope is detecting various amounts of peak to peak voltage with probes sitting on the desk. (Or even completely disconnected from picoscope) I am seeing approx. 80mV P-P when on the 5V scale, up to approx. 350mV on the 20V scale. I am trying to see if the output has less than 60mV ripple but I am above that before even starting. I have tried to attach a screenshot of this. It shows that the Measured DC average is in uV ( disconnected, probably picking up induction) and the peak to peak voltage is 176mV. I have the probes set to 1x in the software and on the physical probe. I am lost what to try next.
Attachments
Capture.PNG

bennog
Advanced User
Advanced User
Posts: 208
Joined: Mon Nov 26, 2012 9:16 am
Location: Netherlands

Re: Peak to peak voltage errors

Post by bennog »

Remember it is a 8 bit ADC scope, so at 20V = +/- 20 = 40V total range.
40V / 256 = 0.156V = 156mv so wat you see is totally correct for a 8 bit scope.

If you want to measure ripple you need to use the scope in AC mode not in DC mode, then you can choose e.g a 5mV range of the scope.

Benno