Using calibrated PT100 probes with PT-104 logger

Post any questions you may have about our current range of USB data loggers
Post Reply
mcameron
Newbie
Posts: 0
Joined: Tue Sep 25, 2018 1:35 pm

Using calibrated PT100 probes with PT-104 logger

Post by mcameron » Tue Sep 25, 2018 1:41 pm

I recently sent away a PT104 and 4 PT100 probes for calibration. The calibration sheets that were returned contain the applied temperature, the temperature that the pt-104 measured and the temperature error.

Can anyone tell me how i can get this info into the picolog software so that i can start measuring using the calibration data?

thanks in advance

Mike

Martyn
Site Admin
Site Admin
Posts: 3799
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Using calibrated PT100 probes with PT-104 logger

Post by Martyn » Fri Sep 28, 2018 9:58 am

You will need to apply channel scaling using the values from the calibration report in a table.
Martyn
Technical Support Manager

Chrllyd
Newbie
Posts: 0
Joined: Tue Nov 13, 2018 7:15 pm

Re: Using calibrated PT100 probes with PT-104 logger

Post by Chrllyd » Tue Nov 13, 2018 7:20 pm

Hi,

Could you please confirm the format of the table. I can see the example in the PicoLog Self-Help Guide:

aa bb
cc dd

where the first column is the indicated reading, and the second is the true reading. Which column should the reference temperature be in (the actual temperature that we are trying to adjust the thermometers to read)?

Thank you!

Post Reply