Post by silhas »

Hello,

I've got some problems regarding plugging a loadcell into my data logger ADC-24.

The current situation:
I got a voltage supply of 24V, a data loader ADC-24 and a force transducer S9M. The force transducer requires a supply voltage between 0.5V to 12V, so I divide the voltage using 2 resistors of equal caliber (one serial connected and one prallel connected).
After the supply voltage is plugged in to the force transducer I connect the (+) measurement signal from the force transducer into the Data logger. The (-) measurement signal i plug into the AG.

The problem:
For starters I only get 4V over the force transducer. This can depend on the load, since the signal is 2 mV/V and thus the main voltage should change as well.
But the real problem is that i get 1960mV from the measurement signal inte the Data logger! This is way above the correct voltage, and I'm in a quite standstil on what to do...

wiring diagram and S9M datasheet in attachment.
Attachments
wiring diagram
wiring diagram
7844_download.pdf
S9M data scheet
(224.69 KiB) Downloaded 272 times

Martyn
Site Admin
Site Admin
Posts: 3833
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Loadcell S9M 2kN into Data logger ADC-24

Post by Martyn »

I am not sure that the "Measurement -" and "Excitation -" should be connected together, surely that will short out one side of the bridge as shown in the datasheet. Additionally the supply needs to be between 0.5V and 12V so I would suggest changing the divider so that it is within the range not at one end.
Martyn
Technical Support Manager

silhas
Newbie
Posts: 0
Joined: Tue Sep 24, 2013 9:02 am

Re: Loadcell S9M 2kN into Data logger ADC-24

Post by silhas »

Well, you are probably correct so I did some modifications to the wiring diagram.

I use a voltage regulator to lower the voltage to 5V, then I use the differential measurement to remove the common voltage. At zero load I get 0 voltage (which is correct), and I then get an increase in voltage as I increase the load. Unfortunately I get a lot of fluktuations in my measurements, and when I change +- voltole range I get completely different measurements. I was wondering why I get theese different measurements depending on my range?
Sometimes when I change the range to a different value and then change back I get a different value from the first time on the same range value even tho I dont change my load!

Is this due to incorrect settings or incorrect soddering?
Attachments
Loadcell with attached voltage regulator
Loadcell with attached voltage regulator

Martyn
Site Admin
Site Admin
Posts: 3833
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Loadcell S9M 2kN into Data logger ADC-24

Post by Martyn »

Are you using your own program ? if so can you post it here.
Martyn
Technical Support Manager

silhas
Newbie
Posts: 0
Joined: Tue Sep 24, 2013 9:02 am

Re: Loadcell S9M 2kN into Data logger ADC-24

Post by silhas »

Nope, I use Picologg Recorder.
The link shows some photos of the wiring and printscreens from the program. You can see the description (voltage range and current load) in the title of the images. The channel currently in use is channel 15 (differential measurement between 15 and 16).

http://imgur.com/a/pYjoG