Post by MaNiaC.at » Sat Nov 26, 2011 9:12 am

Hello,

Apologies if this question has been answered before, or if I could have found the answer myself by studying the documentation more closely!

I'm having the PicoScope 2205. It's maximum samplerate is 200MS/s --> 5ns. In the ps2000pg.en.pdf is the TIMEBASE describe as a code between 0 and the maximum timebase a vailable (consult the driver header file). Timebase 0 gives the maximum sample rate aviable, timebase 1 selects a sample rate half as fast.

So i download the PS2000sdk_r10_3_0_3 and start the excel file, then de VBA. i change timebase in the GetData-funktion from default 13 to 2, the sample intervall was 20ns. I change to 1, the sample intervall change to 10ns. But if i change the timebase to 0 (as in the .pdf described) the VBA didn't worked any more.
I need the 5ns sampling rate. Please help me.

thank you

Michael

Martyn
Site Admin
Site Admin
Posts: 3412
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Sampling rate at the PS2205

Post by Martyn » Mon Nov 28, 2011 10:55 am

If you have two channels enabled then the maximum sample rate is 100MS/s giving 10ns sample interval
Martyn
Technical Support Manager

MaNiaC.at
Newbie
Posts: 0
Joined: Fri Nov 25, 2011 3:04 pm

Re: Sampling rate at the PS2205

Post by MaNiaC.at » Mon Nov 28, 2011 12:57 pm

Hello Martyn

Thank you very much.
I thought that Channel B is default inactiv.
Now i set it to inactiv and it works

Thanks again.

Michael