Maximum voltage/current that can be drawn from signal gen?

Post any questions you may have about our current range of oscilloscopes
Post Reply
furuian
Newbie
Posts: 0
Joined: Wed May 11, 2016 2:02 pm

Maximum voltage/current that can be drawn from signal gen?

Post by furuian »

Hello,

In our lab we have a 5444A and a 5444B. I use the signal generator to generate simple sine waves that we connect to piezoelectric ultrasound transducers. The transducer impedances are between 20 and 500 Ohm (depending on the frequency).

I would like to maximize the power provided to the transducers. What voltage/current/power limits do I have to be aware of in order to not damage the signal generator?

Thank you,
Ian

Arbitrageur
Newbie
Posts: 0
Joined: Sat Jan 14, 2017 11:10 pm

Re: Maximum voltage/current that can be drawn from signal ge

Post by Arbitrageur »

Hi,
I would like to know the official answer as well. On the data sheet tough I found that the output resistance of my 3306D's generator output is 600Ohm. Taking the maximum of 2V output, it means even if I short it to the ground the current is I=U/R 3.3mA.

In essence I do not think you can damage the unit with high amps, since even short circuit operates at very low amps. On the other hand it has overvoltage protection only below 20V, so you should avoid mistakenly putting voltage higher than that on the output.

Best regard,
Justin

Martyn
Site Admin
Site Admin
Posts: 4491
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Maximum voltage/current that can be drawn from signal ge

Post by Martyn »

That is correct for the 3000 series.

Similar for the 5000 series, except the output impedance is 50 Ω, so 40mA.

Both have overvoltage protect at ±20 V
Martyn
Technical Support Manager

Post Reply