How fast can you switch between input ranges?

Which product is right for your exact requirements
Post Reply
rvm
User
User
Posts: 2
Joined: Wed Aug 21, 2024 2:39 pm

How fast can you switch between input ranges?

Post by rvm »

I am considering to buy a Picoscope 4824A, but I am not entirely sure whether this scope is suitable for my application.

I have three coils [A, B and C] that are periodically (at a known time) switching from a 2V signal to a 50mV signal. I would like to measure both signals at the highest possible resolution, so I see two options:
1. I connect each coil to two inputs, one with a 2V input range and one with a 50mV input range. QUESTION: Can the picoscope handle a 2V input signal, while the input is configured with a 50mV input range? Of course this will result in a useless measurement, but will it also break the hardware?
2. I connect each coil to one input and fast switch between the 2V and 50mV input range. QUESTION: can I reconfigure the picoscope from a 2V to 50mV input range within 1millisecond?

Many thanks in advance!
Gerry
PICO STAFF
PICO STAFF
Posts: 1169
Joined: Mon Aug 11, 2014 11:14 am

Re: How fast can you switch between input ranges?

Post by Gerry »

Hi rvm,

I'm not sure if you are still looking for an answer to this, but if you are, then the answer to your question is that (1) would be OK. A ±2V, 500Hz waveform continuously applied to the ±50mV Input Range of the PS4824A would only cause overranging alerts, and is not going to damage the PicoScope over time.

Regards,

Gerry
Gerry
Technical Specialist
rvm
User
User
Posts: 2
Joined: Wed Aug 21, 2024 2:39 pm

Re: How fast can you switch between input ranges?

Post by rvm »

Thank you Gerry. My second question is not relevant anymore. So this answers my question.