ADC-20 Input Impedance

Post any questions you may have about our current range of USB data loggers
Post Reply
marko
Newbie
Posts: 1
Joined: Wed Jun 10, 2009 2:02 pm

ADC-20 Input Impedance

Post by marko »

I would like to use the ADC-20 to make low-level current measurements (nA...uA) by measuring the voltage drop across a shunt resistor. The shunt is likely to have such a large value that the input impedance of ADC-20 may affect the measurement.

The input impedance of ADC-20 is specified to be 1 Mohm. How accurate is this? Can the excact value be checked by a multimeter?

As I am measuring a current into earth, could one use the input impedance directly as a shunt? With 1 Mohm one would get a scale of 1V/1uA. Is this realistic?

PeterF
Advanced User
Advanced User
Posts: 435
Joined: Thu Jun 07, 2007 10:53 am
Location: Cambridgeshire

Post by PeterF »

Hi,
The input impedance of the ADC-20 is 1Meg ohm in single ended mode and is stable. I do not have a figure for the absolute accuracy but it would be easy to measure it. If you have an accurate 1uA current source you could inject it into the input and measure the resulting voltage. Any errors could be compensated for and the correct units displayed by using the "Scaling" option in channel set-up.
PeterF.