TC-08 USB Isolation

Post by nconrad »

Hi,

In our application, we'd like isolation between the TC-08's ground and our thermocouples.

Does the TC-08 provide any sort of isolation already? I noticed that its USB connector's color is orange, and I was wondering if that implied that it was isolated?

Thanks,

Nathan

Gerry
PICO STAFF
PICO STAFF
Posts: 1145
Joined: Mon Aug 11, 2014 11:14 am

Re: TC-08 USB Isolation

Post by Gerry »

Hi Nathan,

The USB connector on the TC-08 is just a standard USB connector (the colour has no significance regarding the USB implemented inside the TC-08). So, the TC-08 doesn't have isolated grounds at the input.

I'm assuming that you'd like to perform the temperature measurement with the Thermocouple in contact with the surface that needs to be measured (which is why you need the isolation) in which case, could you tell us exactly what you will be logging the temperature of, and what temperature you will be measuring to.

Regards,

Gerry
Gerry
Technical Specialist

nconrad
Newbie
Posts: 0
Joined: Sun Jan 21, 2018 8:24 pm

Re: TC-08 USB Isolation

Post by nconrad »

Thanks for the reply.

We're measuring temperatures between 10 and 55 C. The thermocouple will hopefully not come in contact with a signal, but it's possible that it will (on accident).

The thermocouple will be wired along the path following wiring with a RF signal (60 Vpeak), and we've demonstrated good rejection of that signal with the TC-08 logger. Mechanical damage could cause these wires to short. I'm planning on buying a USB isolator in order to remove the risk of damage to the TC-08.

I'd rather use a RTD, but due to size/mechanical restrictions our mechanical engineering folks prefer thermocouples.

We actually only need a single thermocouple channel, but due to good results with the TC-08, we'd like to continue using it.
Last edited by nconrad on Thu Feb 15, 2018 1:57 am, edited 1 time in total.

nconrad
Newbie
Posts: 0
Joined: Sun Jan 21, 2018 8:24 pm

Re: TC-08 USB Isolation

Post by nconrad »

I forgot to mention that it's also possible that our thermocouple will come in contact with salt water or conductive gels, but those are hopefully close to ground potential so it would be more of a measurement accuracy concern than a equipment damage concern. I believe our previous testing was done using a laptop, running on battery power (isolated).

Gerry
PICO STAFF
PICO STAFF
Posts: 1145
Joined: Mon Aug 11, 2014 11:14 am

Re: TC-08 USB Isolation

Post by Gerry »

Hi Nathan,

OK, an isolator would be the way to go then, and you will obviously need one that breaks ground loops. Fortunately, you would be able to use a Full-Speed USB 2.0 isolator Like this one (http://cds.linear.com/docs/en/product-s ... d/2884.pdf) (High-Speed USB isolators that work with our High-Speed data acquisition products are harder to find).

As an aside, I would also check the potential corrosion due to salt water, based upon the alloy used for the thermocouple type.

Regards,

Gerry
Gerry
Technical Specialist