Post by Bef » Thu Jan 10, 2019 7:59 am

Hello,
I'm using Ethernet UDP "raw" commands to set up a PT104 and to read data.

That works perfectly. (Lock device---read EEPROM-data---get measurement data---calculate resistance--- calculate temperature)
Now I'm wondering how to set up different PT100 wire configurations.
In the PICO-Software there are options for 2-wire, 3-wire, 4-wire, etc.
In the programmer's guide I only see 5 commands (0x30 - 0x34). The only configuration I see is the activation and the gain setting for each channel.

Are there any more commands? Or do I have to set up the PT104 via USB?

Thank you for your help!

Martyn
Site Admin
Site Admin
Posts: 3635
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: PT104 Settings with UDP

Post by Martyn » Fri Jan 11, 2019 7:54 am

The number of wires does not change what happens in the device, it just changes the way you use the readings that are returned. This calculation is handled in the driver

For 2 and 4 wire
result = (channel calibration * (measurement 3 - measurement 2))/(measurement 1 - measurement 0)

For 3 wire
result = (channel calibration * ((measurement 3 - ( measurement 2 - measurement 1)) - measurement 2))/(measurement 1 - measurement 0)