ADC11/12 Fast Logging

Postby Guest » Fri Mar 05, 2004 10:41 am

Can you have multiple files for one run from the logger. I have a problem of logging at a fast rate (1/15th sec for 10 channels) and the file size is so big 1,000,000 readings Exell will only import data untill it is full ie 64,000 lines and the rest of the data cannot be accessed.

JKB
Guest
 

Postby markspencer » Mon Mar 08, 2004 9:16 am

Hi,

I am assuming that you are collecting the data with Picolog in 'Real Time' continuous mode. The number of sample in this mode can be set. Reduce this to 64,000 or below and set the data record to repeat immediately.

These can be set when do ing File | New Settings, or through the Settings | Recording for the 'repeat immediately' option and Settings | Sample for 'Maximum number of samples'

Best regards,
Regards,

Mark Spencer