Test and Measurement Forum

Pico 1216 with 5000kg load cell in compression

Post any questions you may have about our current range of USB data loggers

Pico 1216 with 5000kg load cell in compression

Postby trendmarine » Thu Aug 26, 2010 3:48 pm

We're trying to use a Pico 1216 to record the compressive loads on a LCM STA-4 5000kg load cell.

We are exciting the load cell with a stable dedicated 10V supply, and normally the output voltage is reasonably linear, and yields the following (according to its calibration certificate):
0V = 0kg
-1.99mV = 500kg
-9.95mV = 2500kg
-19.91mV = 5000kg

But we get a lot of noise on the zero reading, and no matter how much load we place on the cell, we cannot get a reading to register.
We cannot find a setting to configure the Pico 1216 to accept or record negative voltage. Is it capable of this, or is there another trick we can use to spoof it?
Posts: 0
Joined: Thu Aug 26, 2010 3:25 pm

Re: Pico 1216 with 5000kg load cell in compression

Postby ziko » Fri Aug 27, 2010 10:55 am

Hi the Picolog 1216 measures only 0 to 2.5V you have two options really:

1) You create a circuit that makes it positive and amplifies the signal
2) You use the ADC20 which would work directly.


Kind regards

Technical Specialist
Zen Master
Zen Master
Posts: 1705
Joined: Fri Dec 01, 2006 10:03 am
Location: St Neots

Return to USB PC Data Loggers

Who is online

Users browsing this forum: No registered users and 0 guests