Voltage/Current Datalogging

Post by Mikegr »

Need to be able to monitor at least 2 current channels (0-20A) and one voltage channel (0-800Vdc/AC). Witch Pico Device/Attachments with Cal Certs would be best suited for this application?

PeterF
Advanced User
Advanced User
Posts: 435
Joined: Thu Jun 07, 2007 10:53 am
Location: Cambridgeshire

Re: Voltage/Current Datalogging

Post by PeterF »

Hi,
You can use two "PP264" 20A/60A AC/DC current clamps for the current and a TA057 active differential probe for the high voltage. We cannot calibrate them ourselves so you will need to have them independently calibrated.
You can use a four channel 3424 or 4424, 12 bit scope for which we do offer calibration.
PeterF.