Postby Guest » Wed Oct 22, 2008 7:47 am

I am using accelerometer and load cell as my project. How to set up the sensitivity of the accelrometer and load cell in picoscope software. I have picoscope 3424 and 3224. The sensitivity for the accelerometer is 10/31.6/100mV/g and for the load cell 10.40mV/Lb.F. Thank you
Guest
 

Postby Robin » Wed Oct 22, 2008 1:33 pm

Hi

You can set up custom probes in PicoScope 6. These allow you to specify scaling either by a linear equation or a look-up table. Select "custom probes" from the "tools" menu. Click on "new probe" and this will start a wizard to guide you through the set-up.

I hope this helps

Robin