Channel: Input unit cannot change

Forum for discussing PicoScope version 6 (non-automotive version)
Post Reply
cchuha
Newbie
Posts: 0
Joined: Tue Aug 16, 2022 10:43 pm

Channel: Input unit cannot change

Post by cchuha »

Hi, I inherited a Picoscope 4444 from another project. Looking for some help to change the channel input range from V to A in Picolog6. Is there a menu that I am missing? When I first received the tool, it was in A, after resetting, this is stuck in V. Thanks.
Attachments
Channel Setup Input range.png
Channel Setup Input range.png

bennog
Advanced User
Advanced User
Posts: 146
Joined: Mon Nov 26, 2012 9:16 am
Location: Netherlands

Re: Channel: Input unit cannot change

Post by bennog »

change the probe type to a current clamp or something like that.

You are aware that a scope can only measure voltage on its on and to measure current you measure the voltage over a shunt resistor or use a current clamp (both convert current in voltage in the mV range)

Benno

Martyn
Site Admin
Site Admin
Posts: 4202
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Channel: Input unit cannot change

Post by Martyn »

It is showing that you have a 1:1 Differential Probe connected to the input channel so would normally be measuring Voltage. What you need to do is click on Probe or Scaling to see options, and then chose Equation or Scaling where you can add the conversion to amps related to how you are calculating this. This example is measuring voltage drop across a 1k resistor with return units set to Amps
Attachments
Scaling.png
Martyn
Technical Support Manager