minimum step size of raw adc values

Post any questions you may have about our current range of oscilloscopes
Post Reply
rob_robinson
Newbie
Posts: 0
Joined: Thu May 26, 2022 8:44 am

minimum step size of raw adc values

Post by rob_robinson »

Using a Picoscope 4823 and the python API I have measured and streamed some data of the voltage on a node (see attached for plot). The minimum step size between data points is approximately 19 (sometimes 18 or 20). Why is this the case?

My understanding is that the ADC has a 12 bit resolution.
The ps.ps4000aMaximumValue() return (2^15) -1.

To scale between the two ranges I could understand if the step size was 8 since 2^(15-12) but not why it is 19.

Are there some configuration settings that I am not aware of?

Cheers,
Rob
Attachments
Figure_1.png
Figure_1.png

bennog
Advanced User
Advanced User
Posts: 206
Joined: Mon Nov 26, 2012 9:16 am
Location: Netherlands

Re: minimum step size of raw adc values

Post by bennog »

because 2^(16-12) is 16 and not 8 as you mentioned.

Ant the difference between 19 and 16 is because of calibration correction.

Benno