Calibrate 2208B myself

Post any questions you may have about our current range of oscilloscopes
Post Reply
CommanderLake
Newbie
Posts: 1
Joined: Mon Sep 09, 2019 9:12 pm

Calibrate 2208B myself

Post by CommanderLake »

I know Picoscope has a calibration service to make a bit of extra profit but if they can do it then I can do it too so how do I calibrate it myself assuming I have an accurate voltage reference connected?

G-son
Newbie
Posts: 0
Joined: Wed May 05, 2021 4:11 pm

Re: Calibrate 2208B myself

Post by G-son »

I don't believe you can. You can of course check the scope against a voltage reference, but I don't think you can change what the scope reads.

Martyn
Site Admin
Site Admin
Posts: 4137
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Calibrate 2208B myself

Post by Martyn »

The purpose of a calibration is to check that a device meets the published specifications, and then to provide a table of errors to show exactly where in the specification range that the device currently sits. This allows you to make adjustments to the measurements you make to improve their accuracy, and yes you can do this, and so can a third party calibration house. However if you find that the published specifications are not met then the unit would need to be returned to us for repair and adjustment as this can only be performed by us.
Martyn