I was setting up the TC-08 data logger to collect temperature from 3 channels usin Type T thermocouples.
I connected the first 2 thermocouples the channel 1 and 2 but not channel 3. The logger provided me with temperatures for channel 3 ? whats up.

Gerry
PICO STAFF
PICO STAFF
Posts: 1034
Joined: Mon Aug 11, 2014 11:14 am

Re: TC-08 data logger providing values in channel without thermocouple

Post by Gerry »

Hi Gary,

Looks like you may have a faulty device, could you send your query into our helpdesk via support@picotech.com.

Regards,

Gerry
Gerry
Technical Specialist

Gerry
PICO STAFF
PICO STAFF
Posts: 1034
Joined: Mon Aug 11, 2014 11:14 am

Re: TC-08 data logger providing values in channel without thermocouple

Post by Gerry »

Hi Gary,

I'm reminded of the fact that the TC-08 will show the ambient temperature from measurements of the cold junction temperature for the channel, inside the unit, as soon as the channel is enabled. So, if this is what you were referring to then there is no need to send the device in to us.