Hello PicoScope Support,
I want to compare the the high end time of PWM in a period.
But I can't capture all high end in a period.
As the picture in attachments, capture count is only 1.
So value, min, max, average are always the same.
how can I capture all high end and get min,max value?
Capture count does'n work
Re: Capture count does'n work
Hi TSANYU,
You need to use repeat trigger with multiple buffers of data collected in order to get statistical values of the data. Could you post a psdata file, so that we can see how you have setup the software.
Regards,
Gerry
You need to use repeat trigger with multiple buffers of data collected in order to get statistical values of the data. Could you post a psdata file, so that we can see how you have setup the software.
Regards,
Gerry
Gerry
Technical Specialist