5442D MSO digital probes in PicoLog

Post by tunguskar »

Hey there,

I wanted to activate the digital probes in PicoLog with my 5442D MSO but I'm not able to. How can I activate those channels, I see just the analog probes in PicoLog. I want to trace a SPI connection for a longer periode.

Regards

Martyn
Site Admin
Site Admin
Posts: 4491
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: 5442D MSO digital probes in PicoLog

Post by Martyn »

Digital channels are not included in PicoLog 6 as it only samples at a maximum rate of 1kSample/sec and as most digital buses are faster than this the data would be nonsense.
Martyn
Technical Support Manager

tunguskar
Newbie
Posts: 0
Joined: Wed May 06, 2020 7:56 am

Re: 5442D MSO digital probes in PicoLog

Post by tunguskar »

Is it possible to log the SPI Telegrams with a timestamp and log it life to a given file?