Why the measured value is scaled even I don't set it on PicoScope 6

Post by Link »

Hi!
I'm new to PicoScope 6. I found the measured value showed on the screen is weird.
The device I use is PicoTech 5244D MSO. The software is PicoTech 6.
When I test an 5V output with Channel A, I found the measured value displayed on the screen is 500 mv.The screenshot is attached. I have double-checked the voltage of the tested output is 5 V with multimeter.
So, why the measured value is automatically scaled? Do I have to calibrate first before I use the software?
I also found the measured value is scaled when running the demo provided by you officially from github. So, how to fix this?
Thanks a lot!
Attachments
2.png
1.png

Martyn
Site Admin
Site Admin
Posts: 4500
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Why the measured value is scaled even I don't set it on PicoScope 6

Post by Martyn »

Is the scope probe set on x10 if yes then either change to x1 or use channel options menu to change the probe to x10 in the software.
Martyn
Technical Support Manager

Link
User
User
Posts: 6
Joined: Wed Jul 17, 2019 2:38 pm

Re: Why the measured value is scaled even I don't set it on PicoScope 6

Post by Link »

That's it, I didn't know I should set the probe before. Thanks a lot!