ADC11 connected to two 4-20 mA instruments

Post any questions you may have about our current range of USB data loggers
Post Reply
peterb
User
User
Posts: 6
Joined: Fri Apr 18, 2008 10:28 am
Location: Malvern WORCS UK

ADC11 connected to two 4-20 mA instruments

Post by peterb » Fri Apr 18, 2008 10:46 am

I hope some one can help with this !!
I have been using the ADC with one instrument for some time with perfect results. This circuit is from a battery powered unit so I have had to wire in a 24V DC supply.
The additional instrument is 3 channel and mains powered and when I connect it to the ADC Terminal board all the voltages look correct but when it is plugged into the ADC the 3 new lines go to maximum.

I believe it could be something to do with different ground potentials but have not been able to figure it out!

I have attached a pdf of the circuit to make it easier to understand.

Thanks
Attachments
Contrec 4-20 mA Output.pdf
(49.71 KiB) Downloaded 327 times

PeterF
Advanced User
Advanced User
Posts: 435
Joined: Thu Jun 07, 2007 10:53 am
Location: Cambridgeshire

Post by PeterF » Thu May 22, 2008 2:35 pm

After some measuring of voltages the problem was solved:-

I have checked the various voltages and found the culprit.
�
Measuring between the GND connection of the 120 ohm resistor, and chassis earth I get:
�
From the Contrec Flow Meter,���� 0 Volts
From the ADC Gas Analyser��� ~12 Volts
�
And on the GND pins of the ADC11 when connected to our PC, 0 Volts.
�
It looks like I am going to have to isolate the signals from the Analyser!

Post Reply