TA018 Current Clamp inconsistent

Post by injb »

I'm using a TA018 current clamp probe and I've noticed that I get different readings depending on the angle of the wire going through the clamp. If I twist it a little this way and that, I can get changes of around 20% in the current reading. The clamp is fully closed around the wire all the time

Is that normal? And if so, how do I know which reading to trust? Should I fiddle with it until I get the highest reading?

Thanks.

injb
User
User
Posts: 5
Joined: Thu Nov 08, 2018 1:29 am

Re: TA018 Current Clamp inconsistent

Post by injb »

Tough crowd!

Would it help anyone to answer my question if I provided more info? Would screenshots help?

Thanks!

Martyn
Site Admin
Site Admin
Posts: 4500
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: TA018 Current Clamp inconsistent

Post by Martyn »

The wire will need to go through the centre of the clamp perpendicular to the plane for best accuracy.

What current are you measuring, and is this with a PicoScope? If so a psdata file would be useful to take a look at.
Martyn