Can i use the two channel oscilloscope 5244B with the library?
Just to be sure.

Andreas

Martyn
Site Admin
Site Admin
Posts: 4491
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: 5244B with library?

Post by Martyn »

Yes you can.
Martyn
Technical Support Manager

KAndreas
Newbie
Posts: 0
Joined: Thu Dec 06, 2018 8:43 am

Re: 5244B with library?

Post by KAndreas »

One more question, at the download page i can not find the link to the SDK for the Picoscope 5244D.
Can i user the SDK nevertheless for the 5224D.

I have seen the 5224B is obsolete.

I wan't to measure the raise time of partial discharge impulses. The raise time of this impulses is < 10ns.
In the datasheet of 5244B you meantion a smallest measurable impuls width of 5ns? Could you specify this a little bit more? Can i measure a impulse width of 6,95 ns or will i only see 5ns or 10ns?

Martyn
Site Admin
Site Admin
Posts: 4491
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: 5244B with library?

Post by Martyn »

We have an SDK for the 5000D models although this is currently not available on the website, so please contact support@picotech.com if you require this.

Minimum detectable pulse width 5 ns - this refers to the digital channels on the 5000D MSO models and is not relevant to analog channels.

The rise times for the B and D models are the same
5.8nsecs for 60MHz models
3.5nsecs for 100MHz models (except 16bit mode then 5.8nsec)
1.75nsecs for 200MHz models (except 16bit mode then 5.8nsec)
Martyn
Technical Support Manager