I have a 2204a 1) when clicking on icon to start it takes almost 2 minutes to load. 2) there is noise on both channels if I change it to a higher range say 2 volts the A channel noise is gone . B channel is not affected. I can increase to 20 volts with probes in and grounded or probe not connected does not change B channel. Do no have the filtering option.
I am a new user of an" o scope "
Dennis B
new user questions
Re: new user questions
If you are using Windows 10 then make sure you are fully patched up to date, a recent update will fix the slow start up which was related to a component of the O/S that we use within the application.
Can you post pictures, or better still the psdata file (File->Save->SaveCurrentWaveform), that shows the issue.
Can you post pictures, or better still the psdata file (File->Save->SaveCurrentWaveform), that shows the issue.
Martyn
Technical Support Manager
Technical Support Manager