The arbitrary wave form generator doesn't function correctly below 1Hz and completely stops functioning at 50mHz and below.
The square wave function seems to work all the way to 30mHz but accuracy seems to off quiet a bit around 10%

Martyn
Site Admin
Site Admin
Posts: 4491
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: PICOSCOPE 3404B AWG below 1Hz

Post by Martyn »

If you can post a couple of psdata files, with the setups for both situations, we can then take a closer look.
Martyn
Technical Support Manager

benwoodgate
Newbie
Posts: 0
Joined: Sat Oct 27, 2018 11:24 am

Re: PICOSCOPE 3404B AWG below 1Hz

Post by benwoodgate »

Attachment of AWG going from working at 60mHz and not working at 50mHz
Attachment of Square wave set to 100mHz and off frequency
Attachments
AWG set to 50mHz -no longer working.psdata
(22.78 KiB) Downloaded 334 times
AWG at 60mHz- working.psdata
(164.02 KiB) Downloaded 302 times
20181106-0001.psdata
(104.1 KiB) Downloaded 302 times

AndrewA
PICO STAFF
PICO STAFF
Posts: 400
Joined: Tue Oct 21, 2014 3:07 pm

Re: PICOSCOPE 3404B AWG below 1Hz

Post by AndrewA »

Increase the number of samples to the maximum it will go, using the right arrow, in the 'Arbitrary Waveform Generator' Window.
This will also give the best accuracy for frequency.
This worked for the 'AWG set to 50mHz -no longer working.psdata' file you sent in.
This is an issue with PicoScope 6, it should really use the largest buffer for each model scope.
Regards Andrew