Rise Time specification for 2206B AWG

Post any questions you may have about our current range of oscilloscopes
Post Reply
CuriousCorn
Newbie
Posts: 0
Joined: Fri Oct 19, 2018 11:03 am

Rise Time specification for 2206B AWG

Post by CuriousCorn »

Hi all,

A new owner of a shiny Picoscope 2206B here :-)

I was playing around with the scope and the AWG that is built into it (which runs up to 1 MHz for this model). And I am finding that the rise time (and fall time) of the square wave is around 200 ns! This is surprising. A 1 MHz square wave with that rise time does not look much like a square wave. The rise time is constant over the whole frequency range.

Is this correct? What are the specifications for rise/fall time of the AWG?

- thanks and best regards,
Chris

Martyn
Site Admin
Site Admin
Posts: 4491
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Rise Time specification for 2206B AWG

Post by Martyn »

The output bandwidth of the generator on the 2206B is 1MHz, so you can generate a 1MHz sine wave, but the fastest square wave that can reasonably be generated is 200kHz.
Martyn
Technical Support Manager

picothar
Newbie
Posts: 0
Joined: Mon Oct 22, 2018 6:35 pm

Re: Rise Time specification for 2206B AWG

Post by picothar »

You can also look up the PicoScope 2000 series spcifications here and find that the rise time (10% to 90%) for the 2206B AWG is specified to be less than 120 ns.

You can easily add a rise time measurement in the PicoScope software to verify that the scope is within spec.