TC-08 Temp adjustment

Post any questions about our serial port products
Post Reply
User avatar
DGCYS
User
User
Posts: 3
Joined: Tue Jul 17, 2007 12:17 am
Location: Dom. Rep.

TC-08 Temp adjustment

Post by DGCYS »

It's been a while since i worked with this unit. I seem to recall a way to adjust / calibrate temperatue differences a few degrees up or down.
Like 3 t/c's reading 88, 86, 88 being able to adjust the 86 up to 88.
Can this be done?
Thanks,
David

ziko
Advanced User
Advanced User
Posts: 1705
Joined: Fri Dec 01, 2006 10:03 am
Location: St Neots

Post by ziko »

Hi and thank you for your post.

Using PicoLog you can use a look up table when you select your input channels, you would need to do the following:

Options -> Scaling-> Scaling Method (Look Up Table)

you would then for example put the following:

0 4
2 6
5 9
....
86 88

So you would have the raw value then a space then the scaled value.

Hope this helps.

Kind regards
Ziko

Technical Specialist

User avatar
DGCYS
User
User
Posts: 3
Joined: Tue Jul 17, 2007 12:17 am
Location: Dom. Rep.

Thanks

Post by DGCYS »

That helps but after applying a filtered value of 20 all 6 stabilzed to within 1º of each other. This is very acceptable considering i made the T/C's 60 feet long
Thanks,
DAVID