Post by pcr »

20180220-0001.psdata
Persistence view of 100khz square wave from AWG.
(27.78 KiB) Downloaded 411 times
I attach a psdatafile showing the jitter on a 100kHz square wave out of the AWG which seems excessive to me. It is much better for a sine wave and also for lower frequencies. This is not a problem for me and I post it here to bring it to your attention.

Martyn
Site Admin
Site Admin
Posts: 4499
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: 2205A AWG jitter

Post by Martyn »

The update rate of the signal generator is 1.548 MHz which corresponds exactly to the jitter you see.

With a square wave you are going from a high value to a low value in 1 sample so a 1 sample jitter on the edge would be expected.
Martyn