PicoScope 3405D trigger problem

Post your LabVIEW discussions here
Post Reply
Biturbo27
Newbie
Posts: 0
Joined: Wed Nov 08, 2017 8:51 am

PicoScope 3405D trigger problem

Post by Biturbo27 »

Hello everybody,

I want to trigger Channel A of my PicoScope 3405D at 2V at a range of 10V and a Sinus wave of 20 Hz Input Signal.
That's why I used the 'PicoScope3000aExampleStreaming.vi' and added the 'AdvancedtriggerSettings.vi'.
The threshold has to be in an adc-value, that's why I entered the value of 6554.
But the trigger doesn't work because if I run the program the oscilloscopes immediately begins to run. (See Pictures)

Can anyone please help me?
Attachments
PicoScope3405D Trigger_Frontpanel.PNG
PicoScope3405D Trigger_Code.PNG

Hitesh

Re: PicoScope 3405D trigger problem

Post by Hitesh »

Hi Biturbo27,

In streaming mode, the device will begin to start collecting data irrespective of whether a trigger is set or not.

If the amount of data to be collected will be accommodated by the onboard buffer memory, then you can use a block mode capture, where the device will wait until the trigger condition is met before data is returned.

Regards,

Biturbo27
Newbie
Posts: 0
Joined: Wed Nov 08, 2017 8:51 am

Re: PicoScope 3405D trigger problem

Post by Biturbo27 »

Could you tell me or send me an Example-VI how a block mode capture could look like?

Hitesh

Re: PicoScope 3405D trigger problem

Post by Hitesh »

Hi Biturbo27,

You can obtain block mode examples from the relevant sub-folder from the ps3000a folder in our GitHub repository.

Regards,