Hi!

How can I connect 4-20 mA output pressure transducer to PT-104?

Please suggest any changes needed to accommodate this. Do I need any external power source?

Thanks!
junedfird
Newbie
 
Posts: 0
Joined: Thu Jan 03, 2013 11:40 am

Re: Connecting pressure transducer of 4-20mA output

Postby Martyn » Thu Oct 05, 2017 10:49 am

A little bit of Ohms Law to select a resistor the 4 - 20mA current flows through, so that the voltage remains within the 2.5V range of the PT104, and then you can measure the voltage across the resistor.

If the sensor requires power then you will need to provide this as the PT104 is not setup for this.