sample between buffers

Post by CT9A »

How can I see the complete measurement in 1 window, instead of 31 buffers for example. I have this issue that between buffers data is lost. Example, the signal at the end of the buffer is -1V, the next buffer it starts with -3V.
I want to see the complete measurement. Any tips?

Gerry
PICO STAFF
PICO STAFF
Posts: 1145
Joined: Mon Aug 11, 2014 11:14 am

Re: sample between buffers

Post by Gerry »

Hi CT9A,

What is the sampling frequency that you're using now, and what is the slowest sampling frequency that you can tolerate for the captures?

Regards,

Gerry
Gerry
Technical Specialist

alan
Site Admin
Site Admin
Posts: 195
Joined: Thu May 25, 2006 8:02 am

Re: sample between buffers

Post by alan »

Rather than collect 32 short buffers (eg of 1ms/div) try collecting one longer one (eg 20ms/div). You can then zoom in and see all the detail in one continuous block.