No EXT Trigger source for AWG in PicoScope 3206D

Forum for discussing PicoScope version 6 (non-automotive version)

No EXT Trigger source for AWG in PicoScope 3206D

Postby calvin » Thu Jun 08, 2017 8:43 pm

I am trying to generate a wave that is triggered by an external triggering source (JSR pulsar) connected to the EXT port on my picoscope. I connected the GEN channel to channel B. A wave should be triggered every millisecond (1000kHz), however, the wave is only produced once. I know the triggering works, because If I connect Channel A using real ultrasonic waveform out puts from the JSR pulser things look fine. Upon looking through documentation, I saw that there should be an EXT triggering source option for the signal generator, but there is non for me. So how do I properly trigger a signal generated with a picoscope 3206d?
calvin
Newbie
 
Posts: 0
Joined: Thu Jun 08, 2017 8:32 pm

Re: No EXT Trigger source for AWG in PicoScope 3206D

Postby Martyn » Fri Jun 09, 2017 1:56 pm

I have just tested here and if I apply a signal to the Ext input I can generate the number of cycles of the waveform specified in the signal generator dialog.

Can you post the information from Help->About , and can you also post a psdata file with your setup.
Martyn
Technical Support Manager