humidity detector for 2-cell permeability

Which product is right for your exact requirements
Post Reply
duba0018
Newbie
Posts: 1
Joined: Mon Oct 02, 2006 8:07 pm

humidity detector for 2-cell permeability

Post by duba0018 »

We have a permeability cell and a humidity detector for each side of the cell, but now we need to get those measurements into a computer! Help!

Our humidity detectors output 0 to 5V. The manufacturer suggested to use the "reference" voltage from a data logger as the supply voltage. Do you have any dataloggers that will allow us to do this? Can we some how use the +/- 2.5V loggers to span the 0 to 5V range of the detector?

Thanks for any suggestions or advice.

Michael
Advanced User
Advanced User
Posts: 656
Joined: Thu Jul 07, 2005 12:41 pm
Location: St Neots, Cambridgeshire

Post by Michael »

Hello User,

You can use one of our voltage input data loggers with a potential divider to bring the voltage into the 2.5v range.

The reference voltage is developed for reference. There is a very small amount of current available on this voltage output but it is not designed for driving sensors. YOu will have to check the current requirements of your humidity sensor for suitability.

Best regards,
michael
Michael - Tech Support
Pico Technology
Web Support Forum