PT-104: precision when measuring 0.5 Ohm

Which product is right for your exact requirements

PT-104: precision when measuring 0.5 Ohm

Postby Andrei » Thu Sep 14, 2006 8:41 pm

I want to measure specimens within resistance range approx. 0.5 ... 200 Ohm with precision of 0.25%. It looks like PT-104 is up to the task, but I am not quite sure about the measurement precision in the lower range.

Can you tell me what is the relative precision of resistance measurement at 0.5 Ohm, please?


Postby Michael » Mon Sep 18, 2006 8:27 am

Hello Andrei,

The accuracy of resistance measurement is to within 1 µΩ when using the 0-375Ω range. This will cover the range you wish to measure.

Best regards,
Michael - Tech Support
Pico Technology
Web Support Forum
Zen Master
Zen Master
Posts: 656
Joined: Thu Jul 07, 2005 12:41 pm
Location: St Neots, Cambridgeshire

Postby Andrei » Mon Sep 18, 2006 11:06 am

I am no expert on measurement theory, thus, please correct if I am wrong.

1uOhm looks more like a resolution.

RMS noise on range 375Ohm is 10 ppm, which makes 3.75 milliohm, i.e. accuracy 0.75% when measuring 0.5Ohm.

Thus, if I want to lower RMS noise to 0.25%, I have to average over (0.75/0.25)^2 = 9 measurements, which is not that bad.