Postby Dan Acres » Wed Sep 13, 2006 12:18 pm

Hello, I'm looking at the TC-08 for a high accuracy temperture monitoring system within an enclosure.
I'm not quite clear on what the uncertainty on the TC-08 actually is in high accuracy resolution (0.025K) mode.
According to the online spec the uncertainty in measurement is approx. 1 Kelvin at room temperature. Is this correct?
Dan Acres
 

Postby Sarah » Wed Sep 13, 2006 12:23 pm

Hi Dan

The accuracy of the TC-08 is the sum of ±0.2% and ±0.5C. The first refers to the measurement and the second to the cold junction measurement. The accuracy of the sensor also needs to be taken into account.

Therefore at room temperature the accuracy would be less than 1C.

Best Regards
Sarah
 

Postby Guest » Wed Sep 13, 2006 12:28 pm

Sarah wrote:The first refers to the measurement


In centigrade instead of an absolute measurement?
Guest
 

Postby Sarah » Wed Sep 13, 2006 12:29 pm

Hi

Yes it is with reference to centigrade.

Best Regards
Sarah
 

Postby Dan Acres » Wed Sep 13, 2006 12:31 pm

ok, cheers