Signal Generator will not start in Sweep-mode

Forum for discussing PicoScope (version 5)
Post Reply
ziKarl
User
User
Posts: 7
Joined: Mon May 29, 2006 12:08 pm
Location: Vienna

Signal Generator will not start in Sweep-mode

Post by ziKarl »

When using the PicoScope Software the first time (or emulating this by deleting 'psw.ini' and 'PicoScop 3206 SETTINGS.pss') the Signal Generator will not start when Sweep-Mode is set active (also checkmark 'Repeat Signal').

The reason is the default value of 0.1 'Increment' Field. After changing this field to any value greater than 1.0 the generator will work.

Setting back the increment value back to < 1.0 the signal generator seems locked, no frequency or waveform changing is possible.

Deactivating the Sweep-Mode will also release the lock.

Best regards, Karl

BTW: Is it normal, that low frequency waveform (eg. square, 5 Hz) looks very distorted ?

Michael
Advanced User
Advanced User
Posts: 656
Joined: Thu Jul 07, 2005 12:41 pm
Location: St Neots, Cambridgeshire

Post by Michael »

Hello,

Thank you for pointing this out.
I will pass this first time running issue to the software development team. They will follow this up.

The square wave will start to distort at lower frequencies. This is due to the configuration of the signal generator. The primary purpose of the square wave is to tune attenuated probes. This inlolves using a higher frequency.

Hopes this helps.

Best regards,
Michael
Michael - Tech Support