GROSS ERROR IN D.C. AVERAGE MEASUREMENT IN PICOSCOPE

Forum for discussing PicoScope version 6 (non-automotive version)
Post Reply
Glovisol
Advanced User
Advanced User
Posts: 0
Joined: Sat May 11, 2013 10:03 am
Location: San Marzano Oliveto, Asti, Italy

GROSS ERROR IN D.C. AVERAGE MEASUREMENT IN PICOSCOPE

Post by Glovisol »

In Picoscope, depending on scan time, the D.C. average measurement can be correct, or can show a very big error.

I found this while monitoring 120 Hz PWM with a peak value of 3300 mV, but the same error happens at any frequency and at any level with a given ratio between Frequency and Pulse Duration.

I am enclosing two PC screens showing exactly the same PWM, but taken at different scan times, one showing the correct value and the other a grossly incorrect D.C. value.
Attachments
Picoscope anomaly 2.jpg
Picoscope anomaly 1.jpg

Martyn
Site Admin
Site Admin
Posts: 4499
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: GROSS ERROR IN D.C. AVERAGE MEASUREMENT IN PICOSCOPE

Post by Martyn »

The DC average will only be for the data values drawn on the screen.

In the first trace only part of the third cycle is drawn, so a significant part of the zero level is missing. This results in the reading being higher than would be expected.

The second trace shows more cycles, so the incomplete part of the last cycle has less effect on the overall result.
Martyn
Technical Support Manager