The requirement is to remotely (same building within 25 metres) the air temperature in 4 cold stores.

The approach would be to use the EL005 Data Logger, but what is your recommendation for monitoring the air temperatures. Is it by a EL001, EL041 or other.

Regards,


Ian
IanBeauchamp
Newbie
 
Posts: 1
Joined: Thu Apr 20, 2006 2:01 pm

Postby ricardo » Mon Apr 24, 2006 10:20 am

Hi Ian,

You've got a few options. The EL041 thermocouple solution would be the most economical as this can take readings from 4 thermocouples. The only problem is that we do not recommend using thermocouples longer than 10 meters as the signal can degrade. As your stores can be as far as 25m from each other it may not be a good solution.

The other option is to use 2 or more EL001's. These can take up to 3 thermistors (EL015) each, the thermistors have either 5m or 10m cables.

The third option is to use 4 EL026's. This will also give you humidity readings of each store room.

Best regards,

Ric
Ricardo Paloschi
Sales Engineer
Forum Moderator
User avatar
ricardo
Guru
Guru
 
Posts: 71
Joined: Wed May 07, 2003 9:43 am

Environmental Monitoring

Postby IanBeeauchamp » Wed May 03, 2006 12:07 pm

Ricardo,

Thank you for the advice, much appreciated.