I am using a TC-08 for logging both K-type thermocouples and mV inputs. Before I connect any of the thermocouples, there appears to be a ground isolation problem.

When I connect a 32 mV input the reading is unstable and reads about 24.xx mV.
1 - Disconnecting the AC ground wire, from the instrument that is producing the 32 mV, changes the mV reading but does not correct the problem.
2 - However, if I unplug the notebook pc from the AC line (making it run from it’s battery) the reading goes to 32 mV and is stable. This would not be a bad solution if I was not running a 4 day test.

I would be grateful for any help you could provide.

Steve
Steve

Michael
Advanced User
Advanced User
Posts: 656
Joined: Thu Jul 07, 2005 12:41 pm
Location: St Neots, Cambridgeshire

Post by Michael »

Hello Steve,

Your laptop uses a switch moe PSU with a floating ground. YOu will need to earth the TC08 to the same earth as your test rig. You can do this a couple of ways: Earth the TC08 directly from the serial port(if it's a serial port version) or you can earth the chassis of the laptop.

Simply connect a piece of wire to the ground plane of your test rig and to your PC/TC08.

Best regards,
Michael