Datalogging using two ADC-200

Postby LTurley » Tue Mar 28, 2006 11:50 am

I am trying to do datalogging of 2 signals into two adc-200. The problem I have is achieving a sample rate of 1ms. I can specify a 1ms sample rate but in reality it samples at approx 10ms. Is this correct for this kind of set-up or have I missed a setting somewhere?

Thanks in advance,

Postby Michael » Mon Apr 03, 2006 8:52 am


When you are using more than one unit, the only available sampling modes are realtime continuous modes. This limitation of the software restricts sampling rates to >10ms intervals. In order to achieve the higher sampling rates you are looking for, you will need to write your own software or consider a different PicoScope instrument.

Best regards,
Michael - Tech Support
Pico Technology