TC-08 Temperature offset

Having problems ? let us know the details here
Post Reply
DRussell
User
User
Posts: 2
Joined: Thu Feb 25, 2016 8:13 pm

TC-08 Temperature offset

Post by DRussell »

Hello - I have just received the TC-08 unit. I installed the software and played around with my thermocouples. (type K) All was well, so I setup a test. Now all my thermocouples read approx 6degC for room temperature? I'm sure I must have changed a setting somewhere, but I can't find it?
SE031 thermocouples
I currently have 7 installed.
unplugged temperature reads -130degC aproximately
plugged in approx. 6degC
I have verified the thermocouples in another meter.
if I heat them up, they change appropriately, just with this offset?

AndrewA
PICO STAFF
PICO STAFF
Posts: 394
Joined: Tue Oct 21, 2014 3:07 pm

Re: TC-08 Temperature offset

Post by AndrewA »

Hello,

This problem is almost certainly caused by the thermocouple being used touching something which is at a different potential to the TC-08/PC earth. If this voltage is greater than about +/-7V the channel is pulled out of he common mode range of the TC-08. This causes all channels to read very incorrect values. You must ensure that the metal object you are touching with the exposed end of the thermocouple is earthed correctly and at the same potential as the PC earth. If that cannot be done, the thermocouple must be electrically isolated from the metal surface. Failing that, for short term logging, you can use a laptop running on batteries, that is not connected to mains earth.

The overload can be on one channel, or between channels. The TC08 has an input range of approx 70mV. If more than 5V difference occurs between any of the inputs to the TC08 then incorrect readings will result. Additionally if more than 5V difference occurs between the TC08s ground (also computer ground and usually mains earth) then errors will occur. A typical example where this may occur would be if a thermocouple is used to measure the temperature of the motor on a machine tool. The casing of the motor, although earthed may be at a potential several volts different to that of the earth of the computer.

Plug in the thermocouples one by one to locate which are causing the problems. If possible electrically isolate the thermocouple. If this is not possible, run a wire from the metal outer shell of the TC-08 USB connector to a point near the thermocouple tip. This will earth the area near the thermocouple tip and is often enough to 'pull' the two grounds close enough for normal operation.

Laptops running on mains adaptors can also cause problems. Some laptop power supplies do not have an earth connection and produce large amounts of interference which can be picked up by the sensitive TC08. If you have problems when the laptop is run off it mains adaptor, but not on battery, try earthing the laptop. There is often a screw on the back of the laptop that can be used for this purpose. Alternatively run a wire from the TC08 USB connector shell to mains earth.

Please also see this page about thermocouples-
https://www.picotech.com/library/applic ... ation-note
The low half of the page lists precautions when using thermocouples.
Regards Andrew
Technical Specialist

DRussell
User
User
Posts: 2
Joined: Thu Feb 25, 2016 8:13 pm

Re: TC-08 Temperature offset

Post by DRussell »

thank-you for your response! that is good information to know. I'm not sure that applies in this case as I found some further information.
if we install the software on another computer it is fine. no thermocouples were moved/changed. only the usb was unplugged and plugged into the new computer. (I tried uninstall the software from the suspect computer and reinstall with no success - still readings are off)
both computers were plugged into wall power?

AndrewA
PICO STAFF
PICO STAFF
Posts: 394
Joined: Tue Oct 21, 2014 3:07 pm

Re: TC-08 Temperature offset

Post by AndrewA »

Hi,

Is the PC that's not working a laptop?
Try grounding the USB connector shell to mains earth, with a wire.
Regards Andrew
Technical Specialist