Test and Measurement Forum

how to set sample clock timebase rate

Post any questions you may have about our current range of oscilloscopes

how to set sample clock timebase rate

Postby freemason » Tue Sep 01, 2015 7:06 pm

I am using Picoscope 4824, and have a requirement of setting a sample clock of 6.67MHz, how would i set up this?

i have the following setting on national instruments card and i want to replicate this on picoscope
Device: DAQ::1 (PCMCIA-5102)
Sample Rate (S/s): 6.667M
Record Length: 1.051k
Acquisition Mode: N Samples

Configure acquisition settings.png

Posts: 0
Joined: Fri Aug 14, 2015 10:54 pm

Re: how to set sample clock timebase rate

Postby Martyn » Wed Sep 02, 2015 1:39 pm

The master clock of the 4824 is 80Mhz, and the possible sampling frequencies are related to this value, defined by the equation 80/(n+1) where n is the timebase. So for a timebase of 11 the sampling frequency will be 80/(11+1) or 6.6666666... MHz

So provided this is the frequency you actually require, and not specifcally 6.67MHz, then all you would need to do is select a timebase of 11 in the RunBlock function call.
Technical Specialist
Site Admin
Site Admin
Posts: 2314
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Return to USB PC Oscilloscopes

Who is online

Users browsing this forum: No registered users and 0 guests