Test and Measurement Forum

Voltage & current transient recorder

Which product is right for your exact requirements

Voltage & current transient recorder

Postby steveingram » Fri Sep 16, 2005 1:57 pm

Hi,

I need a datalogger/scope to record voltage and current transients that occur for the trip of a circuit breaker. I need to record a minimum of 1 voltage & 1 current (preferably 3 of each) but the scope needs to trigger from a seperate signal. The system is 415V, 3 phase 50Hz so I'll need attenuator probes and current transformers. The trigger signal would be 240V, so another attenuator probe. Do you have a unit which will trigger the recording of the v & i transients and could be left for a week (say) running on it's own? Sample speed required would be 10kHz min, recording a few cycles (50ms) before and after the trigger? The usb data loggers seem good, but I'm not sure if I can trigger them as described above....

Ta very much
steveingram
 

Postby Michael » Tue Sep 20, 2005 3:19 pm

Hello,

Thank you for your post.

Our range of data acquisition devices are mostly PC based. Therefore to get the kind of results you are looking for, you will need to dedicate a PC to the task alongside our loggers.

As for the number of channels, you would be able to achieve the results you are looking for using only one of our scopes. You would need 2x 4 channel scopes and the 3x differential probes for voltage and 3x current clamps. This would be a less cost effective solution.

Best regards,
Michael
Michael - Tech Support
Pico Technology
Web Support Forum
Michael
Zen Master
Zen Master
 
Posts: 656
Joined: Thu Jul 07, 2005 12:41 pm
Location: St Neots, Cambridgeshire

Voltage and current transient recorder

Postby steveingram » Thu Sep 22, 2005 2:15 pm

Thanks for that,

Which logger would you recommend then? Could you confirm that It has the ability to accept an external trigger signal as well.

Thanks
steveingram
 


Return to Pre-Sales Advice