Hi!
Just received my Picoscope 2204A with probes today. I did a probe compensation and I'm wondering if this is OK?
I mean, there is 19 mV spikes present -- is this normal? 19mV seems a lot to me... or are the supplied probes that inaccurate?
To see what I mean, have a look at the image, where I tried to compesante the supplied probe.

Kind regards!
-
bennog
- Active User

- Posts: 35
- Joined: Mon Nov 26, 2012 9:16 am
- Location: Netherlands
Post
by bennog » Mon Jul 13, 2015 6:12 am
According to https://www.picotech.com/oscilloscope/2 ... ifications the 2204A s a 8 bit scope.
In your image the scope is set to +/- 2V
So 4V / 256 is about 15 mV per bit.
And it is normal voor digital scopes to wobble the LSB.
In PS6 you can do resolution enhangement to 12 for your scope, what this actualy do is increas sampling speed and the average out the samples (so you get a virtual resolution of 12 bits)
Regards,
Benno
-
skyflyer
- Newbie
- Posts: 0
- Joined: Fri Jul 10, 2015 9:08 pm
Post
by skyflyer » Mon Jul 13, 2015 9:22 am
Thanks! So, if I understand this correctly, if I were to define the range from 0 do 1.5V, for example, I would get ~6mV accuracy? Whereas by oversampling to 12bit, I could get to 1mV (theoretically).
I didn't know that the vertical scale affects the probe accuracy (resolution).
-
Martyn
- Site Admin

- Posts: 3467
- Joined: Fri Jun 10, 2011 8:15 am
- Location: St. Neots
Post
by Martyn » Tue Jul 14, 2015 11:05 am
Nearly, but it is related to the vertical ranges you can select, 0 to 1.5V is not a settable vertical range for the device.
+/- 2V which gives 4V peak to peak divided by 256 so approximately 15 mV per bit.
+/- 1V which gives 2V peak to peak divided by 256 so approximately 7.5 mV per bit.
Now if you add resolution enhancement at 12bit, instead of dividing by 256 you divide by 4096
+/- 2V which gives 4V peak to peak divided by 4096 so approximately 976 uV per bit.
+/- 1V which gives 2V peak to peak divided by 4096 so approximately 488 uV per bit.
Resolution enhancement uses a moving average filter so it will affect the rise times on the vertical edges of your signals.
Martyn
Technical Support Manager
-
skyflyer
- Newbie
- Posts: 0
- Joined: Fri Jul 10, 2015 9:08 pm
Post
by skyflyer » Tue Jul 14, 2015 2:36 pm
Martyn wrote:Nearly, but it is related to the vertical ranges you can select, 0 to 1.5V is not a settable vertical range for the device.
Yes, I found that out yesterday, while trying to figure out why I can't set the range from 0 to 50V 
It makes sense now. I also saw that oversampling (resolution enhancement) in this case (square signal) does not give desired results, especially because it affects the rise times so badly.